I am using ADC129 with ESp32 for collecting data of small ECG signal . Unwanted noise is facing of

Atta Rehman , 12-17-2024, 07:35 PM
@ Robert Feranec .
I am using ADC129 with ESp32 for collecting data of small ECG signal .
Unwanted noise is facing of 9.76 HZ .
please suggest me solution to remove or possible reason of this 9.76 HZ noise removal .
i have no frequency on board operating at 9.76HZ .
Atta Rehman , 12-17-2024, 07:37 PM
Sir , As per your experience , Please suggest possible reason of low frequency noise reasons. respectful regard Arehamn
QDrives , 12-17-2024, 10:05 PM
What op-amp are you using?
Atta Rehman , 12-17-2024, 10:10 PM
ADS1299
QDrives , 12-17-2024, 10:20 PM
Not other external op-amp?
Ceramic capacitors (class II) or tantalum/electrolytic?
What about the regulator?
Jitter in the clock?
Atta Rehman , 12-17-2024, 10:43 PM
1. Inverting opamp for noise cancellation within ADC. as shown in figure data sheet figure 73.
Atta Rehman , 12-17-2024, 10:45 PM
2. using ceramic capacitors.
3. LDO's using for analog supply of ADC and digital of controller ESp32.
Atta Rehman , 12-17-2024, 10:46 PM
4. ADC internal clock is using , ADS1299 , 2.048MHz
Atta Rehman , 12-17-2024, 11:02 PM
power Spectral density of data , have harmonics of 9.76HZ
QDrives , 12-17-2024, 11:59 PM
I assume two separate LDO's, one for analog and one for digital?
https://www.youtube.com/watch?v=A9Kvst8hl0M
Atta Rehman , 12-18-2024, 10:11 AM
for digital +3.3V
Atta Rehman , 12-18-2024, 10:12 AM
for analog +/- 2.5V
QDrives , 12-18-2024, 03:44 PM
Do you get your values at 20Hz?
If so, 20 [M?] / 2.048 [MHz] = 9.765625

If you make your picture any smaller, you might as well not post them at all.
I see inductors and ferrite beads on the outputs. Remove them.
https://www.youtube.com/watch?v=HaLMjVkKYMw
https://www.youtube.com/watch?v=U5_61EJlytc
Atta Rehman , 12-18-2024, 04:33 PM
YES , power spectral density Values starts near 20HZ () ,
Atta Rehman , 12-18-2024, 04:34 PM
and repeats harmonics 9.76HZ multiples.
Atta Rehman , 12-18-2024, 04:35 PM
YES , i already removed FB with 0603 ohm jumper 0R resistor . .
Atta Rehman , 12-18-2024, 04:41 PM
" Do you get your values at 20Hz?
If so, 20 [M?] / 2.048 [MHz] = 9.765625 "

yes _ exactly starts from 20HZ and multiples of 9.76HZ .
Atta Rehman , 12-18-2024, 04:42 PM
why it starts from 20Hz. and get harmonics of 9.76Hz , as frequency is 2.048 MHz
QDrives , 12-18-2024, 05:05 PM
The 'frequency' you **see** starts at 40 / 2.048 [MHz] = 19.53125.
What settings do you have set for the SINC filter?
What sample rate (output SINC filter) are you using?
Atta Rehman , 12-18-2024, 07:42 PM
1. Respectfully , X axis is HZ , First Peak starts at 19.53 HZ than 9.76 multiples peaks. it should be in MHZ harmonics peaks are in HZ. Don't under stand _40 digit._
2. ADS1299 has fix **third order SINC** filter **. (not programmable ).
3. i am using **250 SPS** data rate .
Atta Rehman , 12-18-2024, 07:53 PM
ADS _SPI CLK _10MHZ
QDrives , 12-18-2024, 10:19 PM
https://www.youtube.com/watch?v=mCWC0X2naY0
And yes, I know your X axis is in Hz.
Atta Rehman , 12-19-2024, 11:42 PM
thanks dear , I will share data after iteration and debugging update .